my protein gives huge tailing.. - rp-hplc for a protein (Aug/18/2007 )
hii all!
i m doing an rp-hplc for a protein.. on c4 column.. acetonitrile + tfa and water with tfa gradient. with any amounts of protein loaded, my protein gives huge tailing.. wat are the reasons. how do i overcome these?
thanku!!
-alice!-
It looks as though the protein is too sticky for the column. Check the column with other proteins, and see how they behave. Is the protein very hydrophobic?
Q1: How old is the column?
Q2: Are you confident the column is in good condition?
Q3: Do you have any other columns you cold use, say, a C8 or C18? If not, have you tried other solvents, such as isopropanol?
Do you have to use RP-HPLC? HIC can work well with hydrophobic proteins, and you can usually rely on ion exchange.
All the best.
-swanny-