electroporation using Electroporator2510 - Time Const has just 2 msec (Mar/02/2007 )
I m using your Electroporator2510 to make electroporation using E. co//JM109 .pir .
The competent cell was made as usual: harvesting the OD600 0.6-0.8 cells ,  
 ice chilled 15 min,  -1 C  centrifuge for 5 min , then ice cold dd H2O  washing 3 x , 
the final ca.100 microliter rest kept for  electroporation. ( 1mm gap cuvettes, 1350 v , 
Pulse/Pushed 2x , “chg” appeared.) 
 
My question is Why the actual Time Const has just 2 msec  ( rather than usual 
5 msec) ?   When I used other competent cell such as  EPI 300 ( Epicentre inc.)
The Time const is  5 msec. What causes this difference?  
if  DNA too less ?  This " 2msec"  gave  no colony at all   in plates  
How to optimize the conditions to improve the transformation using this instrument?
It s a  basic  question , but  important  now.
Any suggestion is greatly appreciated.\\Have a nice weekend!
Sorry for the late response, In my experience working with HD5 alpha electrocompetent cells, generally I get lower transformation times for ligations, than controls (7msec for controls and 4 for ligations), so I think If you have high Ion concentration you get lower transformation times. But when I get lower times, my cells allways arce. So I do not get any colonies in my plates.
I do electrocompetent cells, harvesting the OD to 0.4 - 0.5, ice chilled 20 min, centrifuge 4ºC 5 min at 6000rpm , then washing with Ice cold water 2x, then washing 1x with glicerol 10%, after discard the supernatant from this step I resuspend in 2ml of glicerol 10%, and aliquote 200 microliter on 1.5 ml eppendorf tubes. 
May be you should try the glicerol step to get longer times for transformation.
Hope that helps.