More than 10 min exposure time in chemiluminescence detection - (Feb/05/2016 )
Hi,
1. Have anyone experienced using auto-exposure setting in Biorad's ChemiDoc XRS and found that it took longer than 10 min but still no results shown?
Does that mean there's nothing on the membrane?
2. I plan to reprobe the same membrane with lower dilution of primary antibody. But I am afraid since the membrane has been dried for more than half an hour in the ChemiDoc's darkroom hood while I was testing the exposure time. Would there any effect on the protein that makes it unable to be reprobed?
Thanks in advance.
I have worked with an XRS before - basically as far as I can tell, the XRS is a terrible system with very low sensitivity. I have had signal that is easily seen on a 10 min exposure of film (and odyssesy systems) that was not seen even with a 30 min exposure on an XRS. So basically there may be signal there, just the system isn't picking it up.
2) The protein will probably be fine, in fact drying a membrane can improve signal in many cases. If you wish to prevent drying, you can sandwich the membrane in cling-wrap or between two plastic sheets.
bob1 on Sun Feb 7 09:08:46 2016 said:
I have worked with an XRS before - basically as far as I can tell, the XRS is a terrible system with very low sensitivity. I have had signal that is easily seen on a 10 min exposure of film (and odyssesy systems) that was not seen even with a 30 min exposure on an XRS. So basically there may be signal there, just the system isn't picking it up.
2) The protein will probably be fine, in fact drying a membrane can improve signal in many cases. If you wish to prevent drying, you can sandwich the membrane in cling-wrap or between two plastic sheets.
There're these few bands that I found at the bottom of the membrane upon 30 second exposure. I know it's not the target protein since the negative control also has the band.
Do you know what could it be?