sequencing results - (Oct/07/2014 )
Hi,
I have 2 sequencing results at attachment. They are different PCR products and was used same primers. For 2 samples after PCR purification with exosapit, below protocol was used:
Big Dye v3.1: 2 ul
Buffer: 1 ul
primer: 1 ul from 3.2 pmol
pcr: 1 ul
water: 5 ul
Sephadex was used.
Sequence machine is 3130. Pop7 and buffers are fresh.
Samples were checked with agarose gel electr. after PCR, and their bands are good and one.
But my sequence results are not very good.
What is your recommendations about this results?
Thanks for your help
Regards
I can't tell much without looking at an electropherogram. The most likely cause is either too much or too little template. I'd vote for too much.
I should also mention that almost all sequencing is done with dramatically less BigDye. You can probably get perfectly good results reducing the BigDye amount to 0.5 ul and increasing the water to compensate. This might save you some money.
What is the ideal template (pcr product) amount for total 10 ul volume?
I would use 10 ng or less in the reaction. Too much is more damaging than too little, typically. Your sequencer manual probably has good advice. It has been a long time since I set up this reaction, so I would read more before taking my advice.
http://openwetware.org/wiki/Optimizing_Sample_for_Sequencing