Equal Loading Confusion - (Jun/09/2013 )
Hi,
Does 'equal loading' on each SDS-PAGE wells means equal sample volume (e.g. 10 ul / lane) or equal amount of protein (e.g. 20 ug / lane)?
Thx.
Usually equal loading of protein content.
but you should also equalize volumes to avoid lane artifacts
bob1 on Sun Jun 9 20:42:56 2013 said:
Usually equal loading of protein content.
Thanks bob1.
mdfenko on Mon Jun 10 11:59:47 2013 said:
but you should also equalize volumes to avoid lane artifacts
mdfenko, do you mean i've to equalize both content (in ug) and volume (in ul)? Then in that case, I would have to dilute my samples to equalize the sample concentrations because my samples have different concentration. For example, from Bradford Assay, Sample A conc. = 2000 ug/ml ; Sample B conc. = 2500 ug/ml ; Sample C conc. = 3000 ug/ml
So do I have to equalize all my samples to the lowest sample concentration, in this case Sample A concentration of 2000 ug/ml? If yes, with what do I dilute my samples?
I've seen protocols which mention that we need to mix our samples with SDS-PAGE sample buffer to at least 1:2 proportion. If i've to equalize all my samples concentration, do I do the equalization before or after mixing the samples with sample buffer at 1:2 proportion?
Thx.
You could do it the way you suggested, or you could take out a certain amount (say 20 ug from each) and then dilute those to the same volume. I would dilute in the lysis buffer, with appropriate amount of sample buffer added.
as bob1 says, dilute with the buffer in which you have the samples (assuming they're all the same). then add the appropriate amount of sds sample buffer for the volume of the diluted samples (this way, all of the samples should be under identical conditions in the gel).
of course, you won't be diluting the whole preparation, just an aliquot with the appropriate amount of protein for the gel.
Thank you so much, bob1 and mdfenko. Your explanations cleared up my confusion now :-)