what does this mean-

"To the solution X was added a solution of 5 eq. of Y in 10 μl THF. The reaction was kept at room temperature for 6 hours. Next, the product was isolated using preparative HPLC to give 15 O.D.A260 of compound".

what is 15 O.D.A260? As far as I know about HPLC, we collect fractions based on the retention time, how do collect fractions based on OD?

-neuron-

Exactly what you said - it is the fraction, they have collected it to give an OD of 15 -so probably an awful lot of protein going through.

-bob1-

(i'm really picking nits here) od at 260 would be for nucleic acids

-mdfenko-

Yes mdfenko, you are right , it is about nucleic acids. @bob, I checked about yields of nucleic acids, they calculate in terms of OD, if you just check synthesis of nucleotides, you will get to know. I am still struggling to calculate concentration, when od is given.

-neuron-

For dsDNA the calculation is quite simple e.g. (copied from cold spring harbour protocols)

Example of Calculation

A sample of dsDNA was diluted 50X. The diluted sample gave a reading of 0.65 on a spectrophotometer at OD260. To determine the concentration of DNA in the original sample, perform the following calculation:
<*>
dsDNA concentration = 50 μg/mL × OD260 × dilution factor
<*>
dsDNA concentration = 50 μg/mL × 0.65 × 50
<*>
dsDNA concentration = 1.63 mg/mL

-bob1-