Protocol Online logo
Top : New Forum Archives (2009-): : SDS-PAGE and Western Blotting

17 kDa ideal transfer voltage - PVDF - (Nov/15/2012 )

Hi all,

Wondering what people would use ideally to transfer a 17kDa protein as efficiently as possibly onto PVDF? I have been doing 1.5 hours at 75V (constant voltage) but am slightly losing the low range. Open for some ideas here. Thanks!

-LacquerHead-

Is that wet transfer? I transfer a 17kDa protein to PVDF by semi-dry at 23V for 1hr. Either way if you're losing the smaller proteins to blow through, then shorten the time or lower the voltage or increase methanol concentration.

-PhDinAcronyms-

Yes wet transfer, I was going to try 75V for an hour? Should be ok I think for the smaller stuff.

-LacquerHead-

75 V for 1 hour is enough to transfer big proteins (100 kDa) across, try about 30 V for an hour.

-bob1-

what is the pore size of the membrane? for low molecular weight proteins you should use 0.22um or smaller.

you can determine if you are blowing your protein through the membrane by using a backup membrane.

-mdfenko-