I am trying to use a standard curve method for qPCR quantification. To do this, I generated PCR products using the same primers that I will use for qPCR using AB's Fast SYBR Green Master Mix and StepOnePlus machine. Then I purified the products using Qiagen's Qiaquick PCR Purification Kit and measured DNA content with nanodrop before making my dilution series of "Standards".

The trouble is that when I generated the standards, the melting temperature for the product was about 85C, however when I use these standards for qPCR the new melting temperature is about 74C, while the melting temperature for my samples is 85C.

Is there an easy explanation for this?

-doxorubicin-

Tm is highly dependent on salt conditions. Are your purified samples dissolved in the same buffers as your test samples?

-phage434-

phage434 on Thu Jul 7 18:46:12 2011 said:


Tm is highly dependent on salt conditions. Are your purified samples dissolved in the same buffers as your test samples?



Both are in TE, which gets dilluted 1:2 in water and then 1:2 again in Master Mix. I am guessing that my standards are too dilute, so there are only primer dimers being formed. I think I need to run some agarose gels to look at the sizes of products etc.

-doxorubicin-

Quick follow-up: Whenever I generate PCR products using Fast SYBR Green Master Mix and then run them out on a gel, they appear as diffuse and "fuzzy" smears. The same PCRs run with NEB OneTaq Master Mix are clean single bands. The qPCR data look nice, but I just don't understand why I am not getting clean PCR products.