DCF Assay to detect oxidative stress - Fluorescence lower than expected - recommended improvements? (Mar/22/2010 )

Hi all,

I am currently trying to use the DCF assay in the flow cytometer. Although the results generally meet my expectations, the changes in fluorescence levels are considerably lower than I expected.

For those of you who are not familiar with the DCF assay, it's a method to measure the levels intracellular reactive oxigen species (ROS). It was first described for microplate reader in 1999 by Wang & Joseph, although the original method had already been discovered in the 60s. Basically, cells are incubated with the profluorescent, lipophilic H2-DCF-DA (dihydrodichlorofluorescein diacetate) which can diffuse through the cell membrane. Inside, the acetate groups are cleaved by cellular esterases so the resulting H2-DCF cannot leave the cells. Reaction with ROS, primarily hydrogen peroxide (H2O2), results in the fluorescent molecule DCF (max. emission ~ 530 nm), so that DCF fluorescence can be used as a measure for intacellular ROS levels.

I've use this method before on a microplate reader with some success, however I would like to try it on a flow cytometer now. I've read some papers on it, of course, so here is the preliminary protocol I've used. Recommendations for improvements are very welcome.

All media contained 10% FBS and 1% Pen/Strep, and all incubation steps were at 37°C and 5% CO2 if not stated otherwise.

- Grow 0.5 x 10^6 cells in 2 mL DMEM for 24 h (cell line Hct-116, adherent human colon carcinoma cells)
- Add testing substance (in this case 50 µM Fisetin) to medium, incubate for another 4 h
- Remove medium, rinse 1x with pre-warmed PBS, add 2 mL pre-warmed DMEM (without phenol red) containing 10 µM H2-DCF-DA; incubate for 30 min
- Remove medium, rinse 1x with pre-warmed PBS, add 2 mL pre-warmed DMEM (without phenol red) containing 250 µM H2O2; incubate for 60 min
- Remove medium, rinse 1x with pre-warmed PBS, harvest cells using 300 µL trypsin for 2-3 min at 37°C, then resuspend in 1 mL ice-cold PBS (Ca/Mg free) and keep on ice in the dark until measurement
- Measure fluorescence in the flow cytometer (Accuri C6, Accuri, Ann Arbor, MI, USA; Ex 488 nm; FL-1 530nm +/- 15; 10,000 events)

The substances I used were Fisetin, a flavonoid that is known to decrease intracellular ROS levels, H2O2 to induce oxidative stress, and the H2-DCF-DA probe. Now the mean DCF fluorescence as represented by the integrated pulse area for FL-1 (530 nm) for each of the six samples was as follows:

(- Fisetin, - H2-DCF-DA, - H2O2) 1838 (1.00) negative control
(+ Fisetin, - H2-DCF-DA, - H2O2) 1991 (1.08) Fisetin autofluorescence control
(- Fisetin, + H2-DCF-DA, - H2O2) 2151 (1.17) DCF / endogenous ROS control
(+ Fisetin, + H2-DCF-DA, - H2O2) 2297 (1.25) Fisetin ROS generation control
(- Fisetin, + H2-DCF-DA, + H2O2) 4674 (2.54) positive control
(+ Fisetin, + H2-DCF-DA, + H2O2) 3508 (1.91) actual sample ;)

Although these results meet my expectations more or less, I would have expected a considerably higher increase in fluorescence levels in the positive control. In similar experiments in the microplate reader, the positive control yielded fluorescence levels up to 20 times those of the DCF control, in which no oxidative stress was induced so that the H2-DCF only reacted with the ROS naturally occuring in the cells. As you can see, here it was only about twice as high, and only about 2.5 times as high as in the negative control, which did not even contain the H2-DCF-DA probe.
Any ideas on what might be the reason for this?

-Aquifex-

Hi Aquifex! Sorry I don't have an answer to your question. I was just looking online for a protocol for DCF assay in a plate reader and your post came out as a result.
Will you be so kind to send me a detailed protocol for that, since you said it worked for you?
Thanks a lot, good luck with your flow cytometry

-manchester-

manchester on Mar 23 2010, 03:22 PM said:

Hi Aquifex! Sorry I don't have an answer to your question. I was just looking online for a protocol for DCF assay in a plate reader and your post came out as a result.
Will you be so kind to send me a detailed protocol for that, since you said it worked for you?
Thanks a lot, good luck with your flow cytometry

Sure, although I'd recommend to check whether you can adapt it to the cells you use. The cell line I use is Hct-116, adherent human colon carcinoma cells. The protocol I used for the plate reader was very similar to the one described above and is actually pretty straightforward:

All media contained 10% FBS and 1% Pen/Strep, and all incubation steps were at 37°C and 5% CO2 if not stated otherwise.

1.) Grow 1 x 10^4 (10,000) cells in 100 µL DMEM per well in a 96 well plate for at least 24 h so they can attach properly
2.) Add in at to medium and incubate for