Low labeling - (Mar/02/2010 )
Hi all,
I'm doing indirect labeling (Aminoallyl + Cy3/5) for a couple of years. I used to get 250-500 pMol of each color with frequency of incorporation (FOI) 10+. Lately I started getting 60-100 pMol with FOI 2-6.
My prorocol is in brief:
1. Amplification with klenow, aminoallyl mix, and random primers ON.
2. The next day I clean with microcon
3. Coupling with the Cy dye in MaHCO3 0.1M buffer for 1.25h. I'm using a fresh buffer each time at pH 8.6.
4. Cleaning with Quiagen purification column.
I tried changing the Aminoallyl stock and the batch of the Cy dye, the klenow enzyme, the random primers. Does anyone have an idea what could be interrupting?
Thanks,
Idit
Hi
Do you cool your samples on ice or on a cooly block after denaturation? A few months ago, my labelling stopped working well and I realised my cooly block kinda broke down and wasn't cold enough! Perhaps it's a long shot. but thought it was worth mentioning
Clare
Idit on Mar 2 2010, 10:31 AM said:
I'm doing indirect labeling (Aminoallyl + Cy3/5) for a couple of years. I used to get 250-500 pMol of each color with frequency of incorporation (FOI) 10+. Lately I started getting 60-100 pMol with FOI 2-6.
My prorocol is in brief:
1. Amplification with klenow, aminoallyl mix, and random primers ON.
2. The next day I clean with microcon
3. Coupling with the Cy dye in MaHCO3 0.1M buffer for 1.25h. I'm using a fresh buffer each time at pH 8.6.
4. Cleaning with Quiagen purification column.
I tried changing the Aminoallyl stock and the batch of the Cy dye, the klenow enzyme, the random primers. Does anyone have an idea what could be interrupting?
Thanks,
Idit
Hi Clare,
Yes, I cool the samples 5 minutes on ice. In what temperature do you do the coupling with the dye?
Thanks,
Idit
Clare on Mar 3 2010, 11:39 AM said:

Do you cool your samples on ice or on a cooly block after denaturation? A few months ago, my labelling stopped working well and I realised my cooly block kinda broke down and wasn't cold enough! Perhaps it's a long shot. but thought it was worth mentioning

Clare
Idit on Mar 2 2010, 10:31 AM said:
I'm doing indirect labeling (Aminoallyl + Cy3/5) for a couple of years. I used to get 250-500 pMol of each color with frequency of incorporation (FOI) 10+. Lately I started getting 60-100 pMol with FOI 2-6.
My prorocol is in brief:
1. Amplification with klenow, aminoallyl mix, and random primers ON.
2. The next day I clean with microcon
3. Coupling with the Cy dye in MaHCO3 0.1M buffer for 1.25h. I'm using a fresh buffer each time at pH 8.6.
4. Cleaning with Quiagen purification column.
I tried changing the Aminoallyl stock and the batch of the Cy dye, the klenow enzyme, the random primers. Does anyone have an idea what could be interrupting?
Thanks,
Idit
Hi again
I use a different method (Invitrogen BioPrime kit) but it's 37degC ON.
Clare
Hi Clare,
Yes, I cool the samples 5 minutes on ice. In what temperature do you do the coupling with the dye?
Thanks,
Idit