connective tissue autofluorescence - (Sep/08/2003 )
My job is to find a way to stain paraffin embedded tissues in Hematoxilin & Eosin, scan the slide (automated machine), destain it and then do Fish stain. That way I can compare the morphology of the cells\tissue it’s cytogenetics. 
After solving almost all of my chemical properties problems (sodium thiocyanate is still missing) I’ve started performing the stains. 
After I destain the H&E I apply DAPI (as for now). I encountered a green background in the fluorescence microscope and I think it’s the connective tissue autofluorescence. 
I read that fixing with formaldehyde may cause autofluorescence with amine groups. 
DAPI has one. So I tried the DAPI stain fixing only with ethanols (not formaldehyde). 
Still, the green background appeared. 
 My guess is that I didn’t destain the eosin good enough.  
Do you know a way of destainin eosin ( pink colored-1% in aqueous solution)? 
Thanks! 
Elad
A long time ago, I read the eosin is fluorescent on the green channel.  I don't think you can circumvent eosin fluorescence and you should avoid it altogether if possible.    
Why not try H&E on one section followed by FISH on the serial section?  Or, do your sections have to be thick enough to contain whole nuclei for FISH?